One way to determine SSDeff is to measure the dose at various distance from the electron
applicator by varying the gap between the phantom surface and the applicator (0 to 15 cm).
At gap = 0, the phantom is in contact with the cone applicator, then the dose is measured in a
phantom at the depth of maximum dose dm. Different measurements are taken with various
distances g from the applicator.1
𝐼𝑜 𝑔 𝐼𝑜
√𝐼𝑔 = 𝑆𝑆𝐷𝑒𝑓𝑓+𝑑𝑚
+ 1 ; √𝐼𝑔 is the slope of the gap distance g
1
Slope = 𝑆𝑆𝐷 𝑒𝑓𝑓+𝑑𝑚
1
SSDeff = – dm
slope
The process
CT planning
The dosimetrist will produce an electron treatment plan with the template for the electron cutout.
The therapist or the staff will make an electron cutout based on the treatment plan.
After the electron cut-out is ready, the dosimetrist or the physicist will take a reading using a
diode to obtain the output factor. The output factor is ratio of the reading with the block and
without the block.
Reference:
1. Strydom W, Parker W, Olivares M. Electron beams: physical and clinical aspects.
Radiation Oncology Physics: a Handbook for Teachers and Students. IAEA. 2005